Bats and shipwreck hunters had the idea first: send out a pulse, time the echo, and you know how far away something is. Do it with megahertz sound from a piezoelectric transducer, which emits the pulses and detects the returning echoes, and you can map a baby without a single ionising photon.
21 original questions · 62 marks · the ultrasound imaging notes · Medical physics
Every question here is written for this library rather than taken from a past paper. Write your answer out before opening a mark scheme: the schemes award marks point by point, and the marks are easier to see when you have something of your own to compare against.
Describe how a piezoelectric transducer both generates and detects ultrasound.
Mark scheme
To generate: an alternating pd applied across the crystal makes it deform and vibrate at the driving frequency, launching an ultrasound pulse (1). To detect: a returning echo deforms the crystal, which produces a measurable pd across it (1).An echo returns 90 μs after a pulse enters soft tissue, where ultrasound travels at 1540 m s⁻¹. Calculate the depth of the reflecting boundary.
Mark scheme
d = ct/2 = (1540 × 90 × 10−6)/2, the factor of two accounting for the there-and-back journey (1)
d = 6.9 × 10−2 m ≈ 6.9 cm (1)State what is meant by the acoustic impedance of a medium, and give its defining equation.
Mark scheme
Acoustic impedance is the product of the density of a medium and the speed of sound in it: Z = ρc (1). Its unit is kg m⁻² s⁻¹ (1).A scanner uses ultrasound of frequency 3.0 MHz. Calculate the wavelength of this ultrasound in soft tissue, where its speed is 1540 m s⁻¹.
Mark scheme
Use of λ = c/f (1). λ = 1540/(3.0 × 106) = 5.1 × 10−4 m = 0.51 mm (1).The acoustic impedance of bone is 7.8 × 10⁶ kg m⁻² s⁻¹ and its density is 1900 kg m⁻³. Calculate the speed of sound in bone.
Mark scheme
Rearranging Z = ρc gives c = Z/ρ = 7.8 × 106/1900 (1) = 4105 m s−1 ≈ 4.1 × 103 m s−1 (1).Calculate the acoustic impedance of liver tissue, of density 1075 kg m⁻³, in which sound travels at 1590 m s⁻¹.
Mark scheme
Z = ρc = 1075 × 1590 (1)
Z = 1.71 × 106 kg m−2 s−1, to the three significant figures the speed of sound carries (1)A boundary lies 3.5 cm below the skin. Calculate the time after the pulse enters the tissue at which its echo returns (speed of sound in tissue 1540 m s⁻¹), and state why the scanner must finish listening before it sends the next pulse.
Mark scheme
t = 2d/c = 2 × 0.035/1540 (1)
t = 45 μs (1)
If a new pulse left before the deepest echoes of the last one arrived, the scanner could not tell which pulse an echo belonged to, and depths would be assigned wrongly (1)Distinguish between an A-scan and a B-scan.
Mark scheme
An A-scan plots echo strength against time along a single line through the body, giving depths along that one direction (1). A B-scan sweeps the beam and turns each echo into a brightness point, building a two-dimensional image of the boundaries (1).Ultrasound meets a boundary between muscle (Z = 1.70 × 10⁶ kg m⁻² s⁻¹) and liver (Z = 1.63 × 10⁶ kg m⁻² s⁻¹). Calculate the fraction of the incident intensity that is reflected.
Explain, in terms of acoustic impedance, why coupling gel is needed between the transducer and the skin.
A scanner sends 5000 pulses each second and must receive all the echoes from one pulse before it sends the next. Show that the greatest depth this scanner can image is about 15 cm. The speed of ultrasound in tissue is 1540 m s⁻¹.
An A-scan, rather than a B-scan, is used to measure the length of a patient's eye before surgery. Suggest why an A-scan is sufficient for this task.
Explain why an MR scan is usually the instrument of choice for imaging soft tissue such as the brain, and state what property of the tissue the returning radio signal depends on.
The acoustic impedance of air is about 430 kg m⁻² s⁻¹ and of soft tissue about 1.63 × 10⁶ kg m⁻² s⁻¹. Calculate the fraction of intensity reflected at an air-tissue boundary, and comment on the result.
In an A-scan of an organ, echoes return 52 μs and 78 μs after the pulse enters the tissue (c = 1540 m s⁻¹). Calculate the depth of each surface of the organ and hence its thickness along the beam.
A student says higher-frequency ultrasound always gives a better scan. Discuss whether this is true.
Give two reasons why ultrasound, rather than X-rays, is used for antenatal scanning, and one limitation of ultrasound imaging.
A sonographer can use one of two coupling gels. Gel P has acoustic impedance 1.50 × 10⁶ kg m⁻² s⁻¹ and gel Q has acoustic impedance 1.10 × 10⁶ kg m⁻² s⁻¹. The acoustic impedance of skin is 1.63 × 10⁶ kg m⁻² s⁻¹. Deduce which gel transmits the greater fraction of the ultrasound intensity into the patient, supporting your answer with calculations.
A scanner converts echo times to depths using a speed of 1540 m s⁻¹ everywhere. A reflecting boundary lies at the bottom of a layer of fat 3.0 cm thick, in which ultrasound actually travels at 1450 m s⁻¹. Determine the depth the scanner displays for this boundary, and the error in the displayed depth.
A B-scan is used to produce an image of a foetus. Describe fully how the image is produced, from the generation of each ultrasound pulse to the formation of the two-dimensional image.
For each of the following patients, choose the most suitable imaging method from MR, CT and ultrasound, giving a reason for each choice: (a) routine monitoring of a foetus; (b) a detailed image of suspected soft-tissue damage inside a knee; (c) urgent imaging of the brain of a patient fitted with an older pacemaker labelled MR Unsafe.
The same practice on paper: the printable workbook for this topic, questions and a worked answer book.
Practise ultrasound imaging one question at a time
The player marks nothing for you. It shows one question, waits, then shows the scheme so you can mark yourself, and brings a question back sooner when it went badly.